Pedro Sousa

Hi!
Almost after 1 month off, I turned the CR on and the battery was empty. After charging it I noticed that the noise (at least in nA range which is the one I use most) was much higher than normal. Actually the open current value is floating from 0 to ~100nA and shorting the output I get roughly the same values and not the expected constant 0 nA. Even more strange, the same happens in the microA range (0-~140 uA). Just in case I reloaded the R3 original firmware and changed the battery but the problem remains. Did I broke the CR? Can it be fixed? The CR (version R3MK4) is less than 1 year old.

Felix

QuoteAlmost after 1 month off, I turned the CR on and the battery was empty.
140 readings with inputs short sounds like a damaged unit.
If battery has some charge before being shelved, it should have been sleeping at <1uA current the whole time, lasting many years.
Battery discharged after 1 month of no use sounds like either a bad battery or the CR was broken before it was turned off, I can't be sure if there's any other explanation why the battery would be drained so fast.
Can it be fixed? Maybe, depends what is wrong. I don't really do repairs, unless there is something really worth learning, you can spend hours debugging what is really wrong. If you can show me the fault is because of a defect in manufacturing or a design fault then I will fix or replace it for free. Design fault does not include user error or not following directions or heed warnings in the guide.

Pedro Sousa


Thanks for the reply, Felix. I will try to check if the battery is damaged. Meanwhile I will contact the European distributor (Welectron) on how to proceed regarding fixing or replacing.